Picoscope 3000s USB lost connectivity

Post any questions related to the new PicoScope 7 software
Post Reply
davetiong
Newbie
Posts: 0
Joined: Wed Aug 07, 2024 5:24 am

Picoscope 3000s USB lost connectivity

Post by davetiong »

hi, i am new to picoscope. i had just purchased 3406D. i am using with teledyne HVP120 probe 100:1 voltage. i need to capture waveform from a DC generator that is able to output 5kW. for a flat line waveform, the picoscope works flawlessly capturing the voltage. problem occurs when i turn on pulsing (square wave) at the generator. even for a 5khz, the picoscope 3406D will just lost connectivity to picoscope 7 or 6. i do notice the green light turns red. the voltage is not at maximum, ~250Vdc. is there anything that i need to configure for the probe or picoscope for it to work correctly?
bennog
Advanced User
Advanced User
Posts: 221
Joined: Mon Nov 26, 2012 9:16 am
Location: Netherlands

Re: Picoscope 3000s USB lost connectivity

Post by bennog »

That type of probe is asking to kill your scope when using that amount of power.
Use a HV differential probe to measure on mains voltage devices.

Benno